Forum Index : Microcontroller and PC projects : TDS/EC Conductivity Meter
Author | Message | ||||
MikeO Senior Member Joined: 11/09/2011 Location: AustraliaPosts: 275 |
The following conductivity meter is based on an original project for Arduino here The original used an American 2 pin power plug as a conductivity sensor, my version is a little different using the ubiquitous "RCA" phono plug, seen here complete with 3D printed holder. Link to STL file You also need a waterproof Temperature sensor (18b20). The Cell is powered through a resistor (470 ohms) from two outputs on the mite via pins ECpower and ECground. Power is applied to the resistor + cell network briefly whilst the junction voltage is measured then turned off, the voltage is then reversed for a brief period to help with water polarisation. The code is fairly well commented and should be self explanatory. Operation is simple and should work straight off but calibration is recommended. I didn't have any commercial standards but was able to make some simple NaCl solutions which seemed to work reasonably well with the aid of this calculator. I made two TDS standards, a 1.5 (1000ppm) and a 3.0 (2000ppm) with 0.29 and 0.58 grams in 250ml of filtered water respectively. Run the software and follow the prompts, if you are going to calibrate the new cell constant may be saved by pressing the S key, you may need to wait out the 10 seconds cycle time before its is saved. The following is a typical Calibrate and Measurement run. New Callibration has been saved Restart in Measure EC/TDS Monitor (Console Mode) Ver 0.6b Defaults - Calibration @: 1.5EC - Temp Coefficient: 0.019 - TDS/ppm Conversion factor: 0.7 Cell Constant: 3.52427 Resistance R1/Ra: 470 / 36ohms - Cell Voltage: 3.28 Vdc Temperature is in Degrees Celcius Waiting for Keypress (C or M) Measure Volt drop:1.031 Temps: 23.25 / 23.25 Resistance: 195.847 Ohms EC:1.50 S Concentration: 1049 ppm/TDS '1.5EC standard Volt drop:1.030 Temps: 23.25 / 23.25 Resistance: 195.671 Ohms EC:1.50 S Concentration: 1050 ppm/TDS 'as used for Volt drop:1.031 Temps: 23.25 / 23.25 Resistance: 195.847 Ohms EC:1.50 S Concentration: 1049 ppm/TDS 'calibration ' Volt drop:2.012 Temps: 23 / 23.25 Resistance: 766.718 Ohms EC:0.38 S Concentration: 269 ppm/TDS 'Filtered Volt drop:2.034 Temps: 23 / 23.25 Resistance: 789.868 Ohms EC:0.37 S Concentration: 261 ppm/TDS 'tap water Volt drop:2.060 Temps: 23 / 23.25 Resistance: 818.633 Ohms EC:0.36 S Concentration: 252 ppm/TDS ' Volt drop:0.741 Temps: 23.25 / 23.25 Resistance: 111.777 Ohms EC:2.63 S Concentration: 1838 ppm/TDS '3.00EC standard Volt drop:0.738 Temps: 23.25 / 23.25 Resistance: 110.809 Ohms EC:2.65 S Concentration: 1854 ppm/TDS Volt drop:0.737 Temps: 23.25 / 23.25 Resistance: 110.533 Ohms EC:2.66 S Concentration: 1859 ppm/TDS Volt drop:0.734 Temps: 23.25 / 23.25 Resistance: 109.982 Ohms EC:2.67 S Concentration: 1868 ppm/TDS > Code for Console version: 'EC Measurment for Micromite 'M Ogden April 2017 'http://www.codenquilts.com.au 'Based on Arduino code by Michael Ratcliffe 'Versions 'v0.5b Console ver Cal, Measure 'Options option explicit option default Integer 'Contants const Sver$="Ver 0.6b" 'Global variables dim float CalibrationEC dim float TempCoef Dim Float PPMconversion dim R1 'resistance of EC Series resistor dim Ra 'resistance of powering pins avg from Microchip data dim float Vin dim float K 'Cell Constant for Measurements, calculated in the Calibration routine dim Float EC dim float EC25 dim float Vdrop dim float Rc Dim ECPin=4 'EC input dim ECGround=17 'EC Ground pin dim ECpower=18 'EC Power pin dim TrPin=16 'Temperature pin Dim RV 'Reverse voltage period value to Cell for Depolarisation Dim TempStart as float dim TempFinish as float Dim Keyin$ Dim ppm,backup,exittask,mode 'Define I/O Setpin ECpin,ain setpin ECGround,dout setpin ECPower,dout 'intialise I/O pin(ECGround)=0 pin(ECPower)=0 'init Modules 'Initialise Arrays ' Restore the settings. If this is the first time set the defaults Var Restore cls Text MM.HRes/2, MM.VRes/2, "INITIALISE", CM, 1,2, RGB(white) pause 1000 '**************************************************************************** 'To restore defaults touch the centre of the LCD "INITIALISE" during startup. '**************************************************************************** if Touch(x) <> -1 Then cls backup=0 Text MM.HRes/2, MM.VRes/2, "DEFAULTS RESTORED", CM, 1, 2, RGB(white) end if pause 1000 If backup = 0 Then backup=1 rv=125 'Reverse voltage period value to Cell for Depolarisation CalibrationEC=1.5 'standard calibration TempCoef=0.019 'temp coefficient for Hydroponics solutions PPMconversion=0.7 'Tranchen [Australia] PPM conversion value: 0.7 R1=470 'resistance of EC Series resistor Ra=36 'resistance of powering pins avg from Microchip data Vin=3.28 'Open circuit voltage reading of Cell K=3.43 'Cell Constant for "RCA type plug" as Cell @ 1.5 EC Var Save backup, CalibrationEC, PPMConversion,TempCoef,R1,Ra,Vin,K EndIf 'interupts 'Settick 1000 ,T1,1 'establish seconds "Tick Timer" ' now draw the display 'init variables R1=R1+Ra mode=0 '0 for Console, 1 for LCD (not implemented) cls '*** start main program loop do consolemode loop '***** End main program loop End If end '**************** sub routines ********************* 'T1 - 1second Tick interrupt 'set up some time flags sub ConsoleMode Do print "" Print "EC/TDS Monitor (Console Mode) ";Sver$ Print "Defaults - Calibration @:";CalibrationEC;"EC - Temp Coefficient:";TempCoef;" - TDS/ppm Conversion factor:";ppmconversion print "Cell Constant K:";k;" Resistance R1/Ra:";r1-Ra;" /";Ra;" ohms";" - Cell Voltage:";Vin;" Vdc" Print "Temperature is in Degrees Celcius" Print "" Print "Waiting for Keypress (C or M)" pause 500 DO keyin$ = INKEY$ LOOP UNTIL keyin$ <> "" 'print "Keyin:";keyin$ If UCase$(keyin$)="C" Then Print "Calibrate, press S (once) to save new Calibration" Calibrate elseif UCase$(keyin$)="M" then ' just measure Print "Measure" Do GetEC printReadings pause 10000 Loop EndIf Loop end sub sub PrintReadings 'Console print "Volt drop:";str$(vdrop,1,3); print " Temps:";tempstart;" / ";tempfinish; print " Resistance:";str$(RC,4,3);" Ohms"; Print " EC:";str$(EC25,1,2);" S"; print " Concentration:";ppm;" ppm/TDS" end Sub sub GetEC 'Read the EC Cell tempstart=tempr(TRpin) 'get start temperature vdrop=EC_Cell() 'read EC cell ' Calculate R relating to Calibration fluid) Rc=(Vdrop*R1)/(Vin-Vdrop) Rc=Rc-Ra EC=1000/(Rc*K) 'compensate for temperature EC25=EC/(1+(tempcoef*(tempstart-25))) ppm=(EC25)*(ppmconversion*1000) end sub function EC_Cell() as float local raw as float EC_Cell=9 'EC can't be 9 so this signals error pin(ECpower)=1 'power on EC probe raw=pin(ecpin) pin(ECPower)=0 'power off EC probe pin(ECGround)=1 'reverse voltage to minimise polarisation pause RV 'pause at reverse polarity pin(ECGround)=0 if raw<>0 then EC_Cell=raw end function Sub Calibrate exittask=0 do tempstart=tempr(TRpin) 'get start temperature vdrop=EC_Cell() 'read EC cell tempfinish=tempr(TRpin) 'get finish temperature 'compensate for temperature EC=calibrationEC*(1+(tempcoef*(tempfinish-25))) ' Calculate R relating to Calibration fluid) Rc=(Vdrop*R1)/(Vin-Vdrop) Rc=Rc-Ra K=1000/(Rc*EC) print "Volt drop:";str$(vdrop,1,3); print " Temps:";tempstart;" / ";tempfinish; print " Resistance:";str$(RC,4,3);" Ohms"; Print " EC:";str$(CalibrationEC,1,2);" S"; print " Cell Constant K:";str$(K,1,2) If tempstart=tempfinish then 'print "Safe to use the K value in main code" else print "Error - Wait for temperature to settle " end if pause 10000 keyin$ = INKEY$ If UCase$(keyin$)="S" Then print "New Callibration has been saved" print "Restart in Measure " var save K exittask=1 end if Loop until exittask=1 End Sub Codenquilts |
||||
TassyJim Guru Joined: 07/08/2011 Location: AustraliaPosts: 5923 |
A very interesting instrument. It looks easy to replicate so thanks for sharing. Jim VK7JH MMedit  MMBasic Help |
||||
Phil23 Guru Joined: 27/03/2016 Location: AustraliaPosts: 1664 |
Very interesting. Continuous pH monitoring is what I'd find really useful, but not sure if the readily available probes are suitable for 24/7 immersion for continuous monitoring. I gather periodical re-calibration might be required, and have seen quite a few continuous pH monitors in service, but they usually have $$$$ price tags. Anyone here know a good starting point for this type of pH monitoring. Thanks Phil. |
||||
paceman Guru Joined: 07/10/2011 Location: AustraliaPosts: 1329 |
You and I and a lot of other people would find that very useful Phil - in my case to control the pool pH. Unfortunately I don't think anyone has yet come up with a simple system that will do this at all reliably. To be able to use pH for control (i.e. not just with colour indicators), pH pretty much universally is measured with a glass electrode which is very delicate and easily and quickly fouled when used for continuous monitoring. Even in pools, which normally have pretty clean water, they foul up fast and at a guess I'd say careful daily calibration with buffer solutions would be necessary. Over longer periods (and not that much longer) they would need cleaning and probably replacement of the electrode solution every week or so. The whole electrode itself won't last long (a few weeks ??) and still be able to give accurate results in a continuous monitoring mode. In the laboratories I've worked in there have been good chemists who understand pH electrodes well, but most would admit having been frustrated by pH electrodes at some stage or other. Greg |
||||
Phil23 Guru Joined: 27/03/2016 Location: AustraliaPosts: 1664 |
Hi Greg, One of the things that spurs my curiosity is a friends Aquaculture setup. An indoor operation of about 1200m², & about 200,000 litres of water. Runs Ozone generators & oxygen concentrators, but the monitoring hardware looks fairly primitive, (Some in service for 15+ years). One of his monitors, runs 24/7, & yep it requires periodical recalibration, but certainly not daily. Gets checked against a higher grade hand held unit most days, but it's pretty low maintenance. More recently he's put in a residential pool & it also has 24/7 ph monitoring & also on demand acid feed to keep pH correct. I know the probes at the fish farm are readily accessible for cleaning etc, but the pool ones are inline & not so readily removed. The couple of devices I've seen so far are these, & I could put both of them to use... It seems to be getting more common for domestic pools to have automatic acid dosing, so there my be a fair degree of reliability available now, given how lax owners can be with correct maintenance. If you could reliably read pH, it would be easy to calculate acid dose required for a pool to maintain the balance. I'd assume it would be introduced slowly, main pump running, and the control would be required not to exceed specified 24 hour doses. Phil. |
||||
palcal Guru Joined: 12/10/2011 Location: AustraliaPosts: 1805 |
PH control would be great.I have another type of meter meant to be used in soil. I bought it in Bunnings (Hardware Store), it has some type of metal probe that must generate a small current because it uses no batteries. It is just the probe with a meter on top it was cheap and is not very accurate but that is not to say it could not be if manufactured properly. Paul. "It is better to be ignorant and ask a stupid question than to be plain Stupid and not ask at all" |
||||
SteveP Newbie Joined: 21/03/2013 Location: United StatesPosts: 19 |
A pH electrode is a voltage source with high impedance output. Generally, a very low bias current op amp is used to measure its voltage, femtoamp is not unheard of. That can be followed by an offset and gain adjustment to convert the small voltage to a pH value. I have seen circuits probably in a low bias current op amp data sheet. pH 0 = +414.12m0V (Acidic). pH 4 = +177.480mV (Acidic). pH 7 = 0.000mV (Neutral). Steve "A chemist" |
||||
paceman Guru Joined: 07/10/2011 Location: AustraliaPosts: 1329 |
Are those monitors DO (dissolved oxygen) or pH? - DO's are quite a lot more robust. The accuracy needed for the pH measurement matters too of course - if the requirements aren't too strict (say +/- 0.2pH units is OK) then calibration can be let out a bit. I once asked our local pool shop how often they calibrated their electrode and they said 'three times a week'. That surprised me a bit because analytically it's done many times a day if the pH measurement is critical - but I shouldn't have been surprised because I'd checked a couple of pool shops against each other a couple of times and they usually didn't match too well. Response speed is another important factor too - in the lab you can't hang around for ten minutes waiting for the electrode to stabilise; 30 seconds would be more likely. Electrodes are thrown out or relegated to less demanding work (pools? ) if their response time blows out. Be interesting to hear his comments about how it's worked - electrode maintenance issue, setup cost etc. I haven't seen any automated pH control on pools around me in Melbourne but then I've only been involved with half a dozen or so and maybe we're backward. Yes I've looked at those too but don't trust the 'reliability' prospect. The meter (i.e. board) should be pretty reliable - it's amazing how far the chip amplifiers have come, and the temp. probe won't be a problem (you have to have temperature to adjust the pH), the electrode is the issue. Those ones are plastic on the outside and that makes them more robust but the membrane inside will still be very thin glass. Those pumps are peristaltic pumps - rotating rollers squeezing a tube. They're commonly used for dosing small quantities intermittently; they're cheap-ish and pretty reliable. Yes, definitely with the filter pump running. I think just using a timing basis for dosing would be the way to go. Run it for a few secs, then off for a minute or so to disperse it, then run again etc. several times for a total of 'n' secs to deliver say 100 mls at a time. The acid should be pre-diluted too I'd say (maybe 10:1?). It takes an hour or two for most pools to circulate the water enough to stabilise pH after dosing any chemicals, so you probably wouldn't want to be dosing acid more than a couple of times a day because it will have to be done during the 'main pump on' filtration periods (2x2 hrs/day in winter and 4x2 in summer for me). Greg |
||||
paceman Guru Joined: 07/10/2011 Location: AustraliaPosts: 1329 |
If it's a pH meter I wonder what the sensor principle is Paul? Soil moisture meters are just a probe with two electrodes, i.e. a rough conductivity meter, but they do need a little power supply, amplifier and some way of reading it. @SteveP Fempto-amps, wow! When I started my training in labs in 1966 a pH meter was real expensive and I guess that's why. Instrumentation amps with that sort of performance weren't common (to say the least). Things have sure come a long way when a cheap Chinese module can do it for you for a few bucks. Greg |
||||
Quazee137 Guru Joined: 07/08/2016 Location: United StatesPosts: 529 |
I have been doing water treatment systems for a few years now and trying to see if I can get the ok to put some of it up here. Till then I can post things that got me to where Im at now as it is open hardware and open source. sparkyswidgets really helped getting some things working right. Lots of great info. I'll have to boot up the old system to get more. I though it was all moved to the new system. Some of the pH probes have been online for at least 2 years with site techs "doing ?" the calibrating and logs. But they do get checked every month by the field rep's as chemical cost eats in to their pay checks. atlas-scientific The PIC based units got me looking and finding MMBasic. I'll add more here as I locate it. |
||||
isochronic Guru Joined: 21/01/2012 Location: AustraliaPosts: 689 |
There was an electrode for approximate pH based on antimony, I think. The potential depended on pH to an extent as it was an oxide/hydrogen ion system. I don't know if it was reliable or not, probably had interference effects from other ions eg chloride. |
||||