Dinosaur![]() Guru ![]() Joined: 12/08/2011 Location: AustraliaPosts: 336 |
Hi All My solar panels when connected to my BM system have a single pair Red / Black. When I measure the current using my Tongs I get a difference of 1.5 A between reading the red or the black.This is in the 16 A range. Even when I measure a load in the mA range the same way, there is always a difference. Granted my Tongs are no-brand, but am I missing something. I am careful about having a nice loop of cable and away from others when I measure. Regards Hervey Bay Qld. |
||||
phil99![]() Guru ![]() Joined: 11/02/2018 Location: AustraliaPosts: 2601 |
If you have the panels connected to a switch-mode MMPT or charge controller the high frequency may be affecting the accuracy of the meter. An old analogue Hot Wire ammeter will give true RMS from DC-to-Daylight. It uses a thermocouple connected to a short piece of resistance wire. Slow response and +/- 3% of full scale are its limitations. |
||||
Dinosaur![]() Guru ![]() Joined: 12/08/2011 Location: AustraliaPosts: 336 |
Hi All My BMS is a Rpi switching SSR's, so the only time the SSR switches is when target Voltage is achieved. I would think the error would be the same on both leads. The tongs actually show the Red lead as Negative current and the Black lead as positive current. Either way whatever comes out of one, MUST go back through the other. So perhaps the accuracy of measurement differs from one direction to the other. Regards Hervey Bay Qld. |
||||
phil99![]() Guru ![]() Joined: 11/02/2018 Location: AustraliaPosts: 2601 |
Turning the meter around at the same place on the same lead should produce the same results if that is the case. You probably need to check it against a known good meter in both directions. |
||||
Dinosaur![]() Guru ![]() Joined: 12/08/2011 Location: AustraliaPosts: 336 |
Hi All Good suggestion, will try that tomorrow. Regards Hervey Bay Qld. |
||||
KeepIS![]() Guru ![]() Joined: 13/10/2014 Location: AustraliaPosts: 1871 |
If by "Tongs" you mean current clamp tester then yes, I find that most DC clamp meters can give a slightly different result in one direction (-/+) and a slight difference between clamping the positive and negative leads. NANO Inverter: Full download - Only Hex Ver 8.1Ks |
||||
![]() |
![]() |